Hi David, I hope you are well? Thank you for the post and thank you Ben too
I guess my question is, are you using PicoLog or PicoScope in the scenario where you have 15 mA before connection?
If PicoScope, Ben has described the zero offset feature that will hopefully resolve the concern and I can now add an attachment to help with interpretation (Attachments were not loading recently)
Given we are measuring tiny voltages; 1.5 mV in the case of your 15 mA “zero value” (0.015 A * 0.1 R = 0.0015 V or 1.5 mV) we need to ensure we do all we can to minimize the impact of environmental noise as just waving the probe in free air with no connection to the vehicle may acquire more than a 1.5 mV in ambient noise.
With channel A options panel open choose the relevant voltage range for your resister.
- Image 1
Remember, approx. 3.1 V drop across the resister would be the maximum for a 100-Watt resister (3.1 V / 0.1 R = 31 A) therefore (3.1 V * 31 A = 96.1 Watts)
Next, select “Zero channel offsets” and join channel A test lead signal wire to ground wire (join 4mm banana connectors together) as instructed below
- Image 2
Select “Continue” and wait for the scope to perform the zero offset. Finally, select Bandwidth limit to “On” to remove any high frequency noise from your measurement process.
You may also choose to lower the sample rate as typically we are do not require a high sample rate to capture parasitic drain.
A final tip is to keep test leads/resister away from sources of noise as this will once again affect your
average parasitic drain value if we are not at true zero
Moving onto PicoLog, if you are using this App then zero offset is not included and so we need to be sure you can zero channel A as described above (using PicoScope)
If zero offset is functioning in PicoScope, we know our measurement hardware and software are functioning correctly. If however after zero offset in PicoScope you still have an offset then we need to investigate further (i.e., try another channel)
Given PicoLog utilizes a very slow sample rate (1 S/s) this has the beneficial side effect of removing noise; note, the same rules still apply regrading keeping test leads/resister away from sources of noise
I am going to raise a New Feature Request for the software team to consider zero offset within PicoLog and hopefully resolve your concern in a future build
I hope this helps, take care…….Steve